Apr 6, 2021 03:16 AM
by WANG
Dear developer of RADIANT SOFT WARE,
I am a fans of RADIANT. I found there is some error in measuring DSA picture. The result of measurenent is larger than the actual size.The measurenent result in RADIANT is the vascular projection on the detetor of DSA. So it is enlarged. The result of measurenent should be corrected by a calibrating ruler. There are some DICOM TAGs including these calibrating infomation. These TAGs are "(0018,1110)Distance Source to Detector" and "(0018,1111)Distance Source to Patient". The corrected measure result is "Not calibrated result"*"Distance Source to Patient"/"Distance Source to Detector".

Thank you very much.

Wang kz
Apr 7, 2021 10:21 AM
by RadiAnt
Thank you for your comments.
Please note that depending on the source of calibration data the length label is displayed in different color: https://www.radiantviewer.com/dicom-viewer-manual/length.html

The only way to obtain exact measurements in the DSA picture is to calibrate length measurements to a known object, e.g., a catheter inside a blood vessel.
The formula that you provided does not account for the thickness of the examined object. There would still be an error between two measurements located on the opposite sides of the examined object.
Post reply
Author (first name or pseudonym):

Message:

Captcha image


Visitors can post messages on the forum without having to register. Please input your first name or pseudonym in the "Author" field. Do not publish any personally identifiable information (PII).

Please ensure you write your posts in English so that our international audience can understand and reply to your message.

The Administrator reserves the right to remove any messages that violate our forum rules; for example, those that are not related to the RadiAnt DICOM Viewer software, published in a language other than English, contravene regulations, etc.